Hi @TaoLiu,
The issue might be related to the object priority settings — Duke should have a higher priority in the voxel setup compared to the electrodes in your current configuration.
From what I can tell, it looks like you’re running an LF Electro Ohmic Quasi-Static simulation (please correct me if that’s not the case, and let us know which solver you’re using). In this setup, it’s generally best to assign the electrodes as boundary conditions rather than as PEC materials, since an object cannot be defined simultaneously as both a material and a boundary condition.
You may also find it helpful to review the “SAM Head with Electrodes” tutorial in the LF section of the manual (Section 3.2.12).